POCT血糖仪与全自动生化分析仪血糖检测结果比对
刘云峰;刘中国;杜利容;邓清华
【期刊名称】《西部医学》 【年(卷),期】2013(025)007
【摘要】Objective To observe the operation of POCT blood glucose monitor and ensure its result accuracy,compared with POCT blood glucose monitor and full automatic biochemical analyzer measurement results.Methods 50 whole blood samples with different blood glucose concentration (low,medium,high) was prepared,and the test results of POCT blood glucose monitor and full automatic biochemical analyzer were compared.Results 7 of 8 POCT blood glucose monitors accord with national requirement after comparing with full automatic biochemical analyzer,while the result error of 1 monitor is larger during measuring the low concentration samples,so its comparison is unqualified because its measured value is above the national standard.Conclusion POCT blood glucose monitor is reliable and convenient,and can be used as emergency,rapid screening examination chemical analyzer for blood glucose detection result is necessary so as to ensure method in clinical application.But regularly comparison of POCT blood glucose monitor and full automatic biothe accuracy of test result.%目的 了解病区POCT血糖仪的使用情况,将POCT血糖仪与实验室全自动生化分析仪血糖检测结果进行比对分析,确保POCT血糖仪检测结果的准确性.方法 配制高、中、低50份不